A question popped up during a recent audit on which I am hoping to gain some of your insight. During a linearity experiment (GC-headspace for acetone) we were questioned over our determination and validity of the "back calculations" stated. In bringing the data, presented below, to the forum I have renamed "back calculation" as "% Recovery". The analyst had calculated the % Recovery at each concentration level versus the theoretical response predicted by the straight line equation, and as the data demonstrates, this recovery falls off at the low end. However, when one calculates the % Recovery versus a predetermined (or "nominal") concentration the data fits nicely throughout the whole experiment.
Internally, we are not unanimous in our choice of the most appropriate method. When dealing with HPLC methodology, I feel that the identification of a "nominal" sample concentration, and thus a reference point for the calculation (cf. the column entitled "From B") makes sense as our work (let's say, an assay) is based around a single-point standard. In the case of GC-headspace, however, everything is quantified by means of a calibration curve. In an equivalent case for headspace I guess this could have been calculated from the "Stock" solution (although it doesn't get injected as it's considered too strong), thus I've used B for the example. As you can see, there is some resulting confusion
As we do not have an "expert" on site, have a QA dept. who know less about it than me and a management barrier-to-development who will no doubt draw out my request for a relevant training course for months, I would sincerely appreciate your advice!

The criterion for our validation is that the % Recovery should be within a certain tolerance, with that value being widened at the low levels. As it stands, it passes by one method and would fail by the other.
Additional criteria are the intercept expressed as a % of the peak area observed at "nominal" concentration. This is another idea brought over from HPLC - it is a confusing concept in GC-headspace due to the aforementioned ambiguity over what constitutes "nominal". Lastly, the % Residual is noted in the final column although I am not sure of it's importance - can it be used to help qualify the assignment of LOQ? At the moment this is done only by demonstrating sufficient precision at low level.
This message has been posted in the LC forum simply due to the greater number of views and the general mathematical topic.