i'm trying to study with igc a natural mineral, a material with high surface energy, using a packed column. My problem is i get unbelievably
short retention times.
i packed a ss column 1/4 inch o.d 20 cm long, i have a normal flow(20 ml/min), the column temprature is 100C and i get retention times about 20 sec for n-hexane and 30 sec for n-decane!
i tried a ss column 1/8 inch o.d 50 cm long with the above conditions and the results wee almost the same.
the last column i made was 1/4 inch o.d 70 cm long and i saw no progress.
i lowered the flow, in 15 ml/min, in all cases but it made no difference.
any ideas?
Posted: Fri Sep 21, 2007 3:31 pm
by Consumer Products Guy
Those are pretty short packed columns. When we used packed columns, 6 feet was a common length, or longer. Need more information.
Posted: Fri Sep 21, 2007 3:54 pm
by stella
Those are pretty short packed columns. When we used packed columns, 6 feet was a common length, or longer. Need more information.
i'll try one about 2m long and we will see. i tried to avoid it because the packing will be difficult and the voids inside the column might be more.
thank you
Posted: Fri Sep 21, 2007 4:17 pm
by MikeD
Stella,
Are you measuring physicochemical properties of your mineral packing ? Can you explain why short retention times are a problem for you.
Posted: Fri Sep 21, 2007 4:28 pm
by stella
Stella,
Are you measuring physicochemical properties of your mineral packing ? Can you explain why short retention times are a problem for you.
i'm trying to determine the dispersive component of the surface energy using apolar probes. The sort retention times are a problem to me for two reasons:
1. a high surface energy material normally gives high retention times
2. as i mentioned the retention times range is very narrow, about 10 seconds from hexane to decane.
Posted: Fri Sep 21, 2007 8:22 pm
by chromatographer1
Just curious..........
what was the retention time for methane?
best wishes,
Rod
Posted: Tue Sep 25, 2007 7:32 am
by Peter Apps
Hi Stella
How big are the particles that you pack into the column ?
Peter
Posted: Wed Sep 26, 2007 8:04 am
by stella
Hi Stella
How big are the particles that you pack into the column ?
Peter
Hi Peter
i thought this parameter might affect the results, so i packed particles with sizes between 100-250 μm, then 100-500 μm and finally 250-500μm (this size is suggested from bibliography, also). None of this three columns worked, the retention times remained short.
Stella
Posted: Wed Sep 26, 2007 8:37 am
by Peter Apps
Hi Stella
How are you measuring the gas flow rate ?
Short retention times could be because the mineral has a low specific surface area - how do the peak shapes look ?
Peter
Posted: Wed Sep 26, 2007 9:22 am
by stella
Hi Stella
How are you measuring the gas flow rate ?
Short retention times could be because the mineral has a low specific surface area - how do the peak shapes look ?
Peter
Dear Peter
i measure the helium flow rate with a soap bubble flowmeter.
the peaks are symmetrical in general with a little tail in their end.
indeed minerals have a low specific surface area, but they have been studied with igc.
Stella
Posted: Wed Sep 26, 2007 12:45 pm
by DR
1) Much longer column - wider bore=easier to pack w/o voids
2) Much narrower particle size distribution - go for as small a particle as you can reproducibly screen without having to screen a dump truck load of it (100-120µ?)
3) Answer Rod's question (Methane RT?) - he has forgotten more about GC than I ever learned

I think he wants to help you establish K. It appears you have had good success separating C6 from C10, so count that as some progress!
Posted: Thu Sep 27, 2007 10:10 am
by MikeD
But if you are measuring a thermodynamic property in inverse GC long packed columns are a confounding factor. The gas compressibility factor can mess up the calculation.