Dear all,
I am sorry about creating a new conversation here while I already posted my problem in "Peak in blank reagent after derivatization". The reason is that I really need the suggestion from all of you. I did contaminate the autosampler with my derivative so even I inject the mobile phase, it still shows a peak but if I connect another loop injector, this peak disappears. HW Mueller and Consumer Products Guy suggested me to clean the autosampler but I don't have any idea how to clean it properly! I flush whole system with IPA and HNO3 already but the contamination is still somewhere in autosampler. Could anyone show me the good procedure to clean the autosampler (including taking it out and cleaning manually)? I really need to get rid of it ASAP because the time for my project almost run out.
Thanks in advance!
-------------------------------------------------------------------------------------------------------
By Consumer Products Guy on Monday, August 23, 2004 - 07:52 am:
First I'd try a new rotor seal (are you using vespel or tefzel?), and flush with solvent stronger than IPA, such as ACN, THF. I physically have removed the stainless steel lines and made up fittings adapters and have flushed from a 10 cc syringe with a luer-lock fitting that has a 1/16 inch needle, to which I swaged HPLC nut/ferrule.
-------------------------------------------------------------------------------------------------------
By anon on Monday, August 23, 2004 - 12:28 pm:
first, i would confirm that it's caused by the ALS. Plumb the system bypassing the ALS, then run several blank gradients to confirm that the peak is not there !. then, change one part at a time to see where the peak goes away....
-------------------------------------------------------------------------------------------------------
By ingochrist on Saturday, August 28, 2004 - 02:13 pm:
What type of autosampler do you use?