- Posts: 46
- Joined: Mon Sep 10, 2007 5:36 pm
I presume the images below are due to column bleed? This is a relatively new 5% phenyl column (Rxi-5MS 30m x 0.25mm x0.25um) with a 1uL splitless injection (purge at 30 seconds) of hexane. The column was conditioned at 340oC for 2 hours prior to use (as per manufacturer instructions). The GC oven ramps at 80 to 320oC over 15 minutes, the injector is set at at 300oC, and transfer line to the MS at 325oC.
TIC

Peak at 7.4 mins