I use SGE Gas Tight syringes for quantitatively dispensing samples that are predominantly volatile organic solvents. I find them really easy to use, once you understand the parallax error, and need for multiple flushes. Also, I don't have to calibrate them as often as pipettors.
I have the range of 25, 50, 100, 250, 500, and 1000 uL. They are PTFE-faced plungers with removeable needles - so I can strip them down and ultrasonicate for cleaning when dealing with cytotoxins. I don't use the luer lock versions. nor the PTFE sealed versions ( too much trouble to clean, and the seals bind after a while with solvents ).
For smaller volumes ( < 10 uL ) I tend to use the 5 and 10 uL guided plunger-in-barrel series, which don't have a PTFE face, but I find the long design allows me to quickly set and dispense the volume. For really small volumes <1 uL, I often use a 0.5 uL or 1 uL plunger-in-needle SGE syringe, but usually would change the dilution to use a larger volume.
In general, I use the same syringe for preparation of standards and samples, so the actual volume conformance specification ( 1% ), is not a major issue. It's important to have several sizes, so graduations are relevant, and samples are near the nominal capacity.
Normally, I quickly flush the syringe with each sample about 10 times ( only takes a few seconds), then slowly draw up the sample past the graduation, and then push down to the mark. I then flush the syringe with sample solvent about 10 times before next sample, unless I know it's a time series experinent, when I ignore the solvent flush.
I quickly wipe the needle and tip with a tissue, and withdraw the plunger a mm or two, however the usual recommendation is to withdraw the plunger and then wipe the needle.
It's really important to quickly, but carefully, wipe the outside of the needle that has contacted the sample, especially if you are making large single dilutions of very small ( < 10 ul ) aliquots. Otherwise you will find calibration curves don't pass through zero.
With regard to the 2% relative error limit of the air displacement pipettor, the specification for the 100 and 1000 uL models I use ( Eppendorf ) is an inaccuracy of +-0.8 ( 100uL) or +-0.6% (1000uL ), and an imprecision of <=0.2%.
As noted earlier, I've never been able to consistently get an acceptable value for use of an air displacement ( tried three different brands ), or even a positive displacement ( Gilson ) pipettor with hexane, or any other volatile, low viscosity, solvent.
If 2% works for you, fine.
Bruce Hamilton