5972 bfb tuning
Posted: Sun Sep 16, 2007 8:29 pm
hello everyone, I am new to this forums and need some help. I run a 5890 connected to a 5972. This particular instrument is giving me some problems at this time I can't pass a BFB tune due to the fact that ion 175 is not present at all in any of the scans, also the baseline is a bit noisy but not to bad.
This is what has resently been done to the instrument, it was taken down and the source was cleaned, both filaments were replaced, the filament cable, power supply, and electron multiplier were also replaced. The top board and side boards were tested and both were found to be doing their respective jobs well. Also the quadrapoles are in good shape. I can get a very good tune of 69 at an abundance of around 1 million, 219 at about 50% of that and 502 at around 2%. With an Em voltage of 1950.