domino1 » Mon Aug 13, 2007 12:47 am
To answer the system peak vs carryover peak...
I just "inherited" this method so I am looking over the work that was previously done on it. The spectra of the peak does look suspiciously like the main peak (although it is pretty ragged due to the low amount).
They looked into washing with 100% ACN between runs, multiple blank injections, changing source of water, TFA, ACN etc and were never able to get rid of the peak (always run on a newer model Alliance). It goes down to a point and then levels off to a constant amount. I am going to try my favorite trick of running a doubled up gradient (to take the injector loop out of the equation) to see if it also shows up in the second half of the chromatogram. Then I throw in a longer equilibration time in the middle which can "concentrate" a mobile phase artifact sometimes. So, I might expect to see a larger peak in the second half of the chromatogram if I double the re-equlibration. If it is gone in the second half of the chromatogram, then my injector loop may be the issue and it is indeed carryover.
The fact that it never goes to zero makes me think it is mobile phase related, however I have a hard time believing that they were not able to get it to go away after changing every single reagent. Plus the spectra point to carryover.
We are going to repeat some of the earlier work (and maybe run on an Agilent) just to verify that there is no way to get rid of this pesky peak before I completely throw up my hands.
Thanks again for everyone's thoughts. It is amusing to me that there are other "geeks" out there who are thinking about HPLC on the weekend.