Celiagg » Wed Sep 12, 2007 10:55 am
Hello again..
I prepare each solution inmediatly before to inject it. Intermediate solutions (in methanol) of standars are at the freezer at -20º C, stored in capsuled vials without head space volume.
Calibration levels are : 1 , 2, 5, 10, 15, 20 ppb, at the aquous solutions. I take 5 mL of them, and inject at the purge-trap system. I purge this for 11 min, and after, i run the program of desorption at temperature and during the time recommended by the supplier of the trap .
I CHANGE THE TRAP AND PUT OTHER KIND OF FILLING and the problem persist ( I had TENAX nº 3 and change to VOCARB 3000).
I've been studing linear range with four calibration curves, and everything was OK. I know someone that works with the same system, and the same stationary phase, but less thikker INOWAX_60m*0.25mmID*0.25 um) and her calibration range is form 1 to 15 ppb without problems.
Yes, cuentas are "counts"; sorry it's Spanish way of speaking.
All the peaks are affected.
Some of them YES: LOOK FLAT.
The BFB (1-Bromo-4-Fluorobenceno) internal standard looks flat in some of the injections.
I've been thinking on it , and if it was a column problem i'd have tailing, and changing in retenction times and it is not so; If it was a saturation in the trap capacity, areas would be higher and are SMALLER.
DO I have leaks in the purge and trap? If it was a leak, I'd have less response, less area, but it would not affect the linear range.
I CAN'NT UNDERSTAND ANYTHING...
Thanks Peter, please, continue helping me...
Kind regards,
Celia