Right now I’m using a 2.1 x 100mm UPLC HSS T3 column, mobile phase is 2 mM ammonium acetate 0.1% FA (A) and Acetonitrile with 0.1% FA (B). If I do a 30:70 isocratic everything comes out in one peak. So I’m trying a gradient starting at 80:20 to 65:35 with flow rate of 0.6 mL/min. This was the gradient my clinical application specialist came up with, I’m afraid I’m quite ignorant when it comes to these gradients as I’m a GC man. All I know is temperature.
Anyway I cannot separate Imipramine from Nortriptyline (and nortriptyline D3) and for some reason the Imipramine peak dwarfs the Nortriptyline peak by a factor of 4 or 5. They (waters) tell me I don’t need separation because it’s MS/MS but from the reading I’ve been doing this is not good because of ion suppression which I think is going on here. So I’m wondering if there is a better column to use. One thing that is irritating is how much pressure it takes to push 80% water at 0.6 mL/min through this column. It is as high as 10,000. I’m afraid if I go to a 150 mm column I’ll go over the 15,000 psi limit.
Another question is I’m extracting my samples with Oasis MCX extraction cartridges. I’m wondering what strength aqueous phase should I reconstitute my samples after drying down the elute (80:20 or 65:35).
Sorry this is so long and hope it makes sense.
TIA