Buffer solution: sodium citrate dihydrate and anhydrous dibasic sodium phosphate, pH 8.
Mobile phase A
(85:15 v/v) Buffer solution : acetonitrile
Mobile phase B
(30:70 v/v) Buffer solution : acetonitrile
Diluent
Dissolve 29.4 g of sodium citrate dihydrate in 1 liter water.
Gradient conditions:
Time
(minutes) MP A MP B
0 100 0
15 50 50
25 0 100
27 100 0
32 100 0
Here is the problem: The active peak on both the Agilent and Alliance systems exhibits the same response, but one of the related compound peaks, which elutes around 28 minutes on both systems, responds differently on the two systems. The Alliance systems give a larger response (2 -10 X) for the related compound peak than the Agilent systems. We have tried this method on multiple systems, columns and with the same sample preps.
Has anyone had a similar experience? We are having trouble figuring out what is going on. Could it be anything with injection system differences? The way the injectors rinse? Dwell volume differences of the two systems?? It's strange that it is only occuring with the one peak and not the others in the chromatogram.
 
 Thanks in advance for any assistance!
Diane














 
																							