eluent containing TFA

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Has any one used TFA in ACN on HPG pump with wavelength 210 nm. I am getting too much noise. A=0.1% TFA water & B= 0.1 % TFA ACN and i am using flat gradient 70:30 (A:B)

Check the absorbance spectrum of each of the mobile phase components, your CH3CN may be absorbing at 210nm. The TFA may also be absorbing, but at 0.1% that usually isn't too much of a problem - providing you are using a suitable grade.

If you want to eliminate pump mixing noise, premix the 70:30 solvent and degas, and use in just one channel.

Bruce Hamilton

The trials and travails of TFA mobile phases have been discussed many times on this forum. Please search the archives. You will find that your problem is very common, and there are things you can do to minimize the noise, but it will never fully go away.
Mark Tracy
Senior Chemist
Dionex Corp.

Lots of information in this article:

http://www.lcgceurope.com/lcgceurope/Th ... tail/45019
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry