Limit of detection via chromatogram [August 11, 2004]

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
By dani andalas on Wednesday, August 11, 2004 - 09:48 pm:

Hello scientist, I have found difficulties how to determine LOD via GC chromatogram. Do I have to inject minimum concentration of standard

-----------------------------------------------------------------------------------
By carla on Sunday, August 15, 2004 - 02:40 pm:

ever tried a statistic software for calibration validation?

It kind of depends on how your company's SOP discussing this is written. As a rule of thumb Ive generally defined LOD as 3 times average noise value and try to get a the LOQ to no greater than 10 x the average noise value. Again, thats just MY OWN personal rule of thumb.
But hey, its an option you may want to consider.
~Law_son
Try using your lowest cal standard as your MDL (method detection limit)
Also, use 2-5 times below your RL (reporting limit). The signal to noise ratio mentioned earlier will work too.
Essentially you want to see how low your instrument can read with 99% confidence that it is > zero.

Semivoadude
3 posts Page 1 of 1

Who is online

In total there are 5 users online :: 2 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: anaamaria1990, Baidu [Spider] and 3 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry