Determining void time for a Atlantis HILIC column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi all,

can anyone tell me the proper way to determine the void time on a HILIC column? I assume sodium nitrate would not be the ideal compound to determine this value on a HILIC column.

Thanks

The easiest way would be to simply inject a blank of ACN/water at a different ratio from your mobile phase, set your detector out at fairly long wavelength and high sensitivity and look for the refractive index shift.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

That would work, though I might do it a bit differently.

I would think that a mobile phase containing anything from 50-95% ACN and a dilute solution of something like toluene in ACN (maybe 200 ppm?) would do nicely enough.

Toluene is what we use for determining the void volume on our Zwitterionic HILIC phases. It works!
Merck SeQuant AB
www.sequant.com
4 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry