Hi,

We have an Agilent LC/MS in the lab, connected to a LC/MS Chemstation Rev. B02.01 - SR2 (260). Recently we come across an error "HED Fault. (0301)" in module G6130A. To get rid of this error, we need to restart the chemstation computer. The error occurs at radom,sometimes even after 30 min after reboot. We do no know what is causing this, nor do the Agilent service engeneers.
Has anyone experienced this problem and can give us some clues how to solve this problem.
Your comments are much appreciated.
Gilbert

I believe this is the High Energy Discharge fault and I think we used to have this because the detector cell inside was contaminated and was caused some stray voltages near the detector. It started to occur after we had a couple of power failures and the oil from the rough pumps made it up into the detector region.

We dissasembled the detector and rinsed everything and I believe this cured it.
2 posts Page 1 of 1

Who is online

In total there are 350 users online :: 2 registered, 0 hidden and 348 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 348 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry