I’ve been have a problem of “driftâ€
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
I have done alternate standard then blank injections to check for carryover and I see no carry-over. Is it possible that the isocratic method allows some of the analyte to stick to the column and remains there until the next sample injection band comes through?Probably carry-over effects from injection to injection...
I am in the field of bioanalytial analysis. It is very typical for us to pre-inject between 10-20 samples to stabilize a LC/MS-MSinstrument before a stable baseline reponse is obtained. In addition to temperature, very often the interfaces have to see enough of the compound of interest for them to start giving stable response. I agree with other posters that use a closely related internal standard is key for normalized reponse ratio.It could be caused by the instability of instrument during its initialization (which supposed to be immediate ideally).
However, the LC-MS interface, ESI or APCI or APPI, which desolvates the LC eluent, needs time to reach the stable temperature. In some case, it requires 45 minutes to reach the temperature plateau.
If what described above is the cause of your observed unstability, give yourself a liitle time to ensure the "initialization" of intrument while you are monitoring the background level.
An easy way to eliminate your problem: use internal standard, stable isotope labeled internal standard as ideal.