Lost of Linearity_P&T-GC-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello Everybody!!

I'm in problems : I'm developing a method and validating it and I've found that I 've lost linear range.....Can someone tell me if it may be a fault in the GC column performance?.Peak shapes are OK. I tried changeing the trap and it does not improve..

Thank you

Celia
Celia

Check your flow rates. Your split ratio can be important with purge and trap (20 mLs/min split vent) and also check your purge and trap sample flow rate (25-30 mLs/min).

celiagg,

Based on your earlier posts, I would propose you reduce your column flow by about 0.4 mL/min. This should put you around 0.8 mL/min which is going to make your run a little longer but I suspect it will improve your linearity. You will want a split ration of about 25:1 to maintain sufficient flow across the trap under these circumstances.

Best regards.
Split flow is very important in the anlysis of volatiles by P&T. Check your injector and liners.
4 posts Page 1 of 1

Who is online

In total there are 216 users online :: 0 registered, 0 hidden and 216 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 216 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry