Hello All,

I am running a reverse phase assay on an agilent 1100 instrument. The mobile phase is a mixture of 50mM Phosphate buffer and Propanol/Acetonitrile.

On some of my runs I have the typical profile with no problems, but it seems that every 4 or 5 runs I get a very unusual baseline. My last inject had some baseline noise every 2 minutes. I then injected again, with no problems. The injection after that had a lot of baseline noise.

Thanks for your help.

Posted: Tue Aug 28, 2007 8:46 pm
by tom jupille
"Baseline noise" covers a lot of ground. Can you be more specific as to what's going on:
- spikes?
- random?
- cyclical?
- lumps (look like they might almost be peaks)?
- drift?

Even better, get a screen shot and show us what the problem (and good) chromatograms look like (instructions are here).

Posted: Tue Aug 28, 2007 10:11 pm
by r6180
Mostly I am seeing small negative peaks throughout the run.

Posted: Wed Aug 29, 2007 4:30 pm
by juddc
Really, a screen shot posted would help quite a bit.

I recently had a baseline issue traced to rapid and excessive humidity changes in the laboratory - 30% swings in RH every 15 minutes drove my PDA crazy. Once I determined that the instrument was fine and was able to quantify the environmental issue (with a Temp / RH data logger), I was able to get our facilities group to fix the problem quickly and my baseline is fine

Posted: Wed Aug 29, 2007 6:56 pm
by tom jupille
Did you do anything between injections? or did the noise gradually taper off during a run?

Also: gradient or isocratic?
what detector? (if UV or PDA, what wavelength).
what injection volume?
what diluent in the sample?

At this point I'll hazard a (very) wild guess that you're seeing small air bubbles building up in the detector cell and then breaking loose.

Posted: Wed Aug 29, 2007 7:36 pm
by r6180
Image[/img]

Here is a picture of the chomatogram.

Some additional Details

Posted: Wed Aug 29, 2007 7:39 pm
by r6180
This is a reverse phase assay. I have tried chaning the column and the mobile phase. Some chromatograms look more typical others have varying degrees of noise.

I am pretty sure that the humidity and temperature in our lab is very stable.

More additional details

Posted: Wed Aug 29, 2007 7:44 pm
by r6180
In between injections I am doing a needle wash. This is an isocratic elution. The mobile phase is an 50mM Phosphate/ Acetonitrile/Propanol mixture.

We are using a UV/VIS detector with the VIS lamp off. Detecting at 214 nm. The injection volume is 10 uL. This is an injection of protein sample in formulation buffer.