I use a Waters 2790 as an autosampler allowing OpenAccess direct infusion to an LCT mass spectrometer.

This has worked well in the past, but recently the 2790 has started to give frequent error messages "Lost prime". Cancelling the error message allows continued operation without any other action. The error message usually occurs at the end of the sample, frequency sometimes approaching every sample!

I've tried a variety of options none of which seem to make much difference. Any thoughts?

Things I've tried.
Wet prime
Dry priming the whole system followed by wet priming.
Changing degasser option, sonicating solvent before running
checking for leak, none found.
Different solvent batches

Thanks.

Might be time for new check valves
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

In the mean time, don't let it stop pumping between samples. Keeping at least a slight flow will discourage the event.
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry