I think that the phospholipids are being retained inside the drift tube. Does anyone have a cleaning procedure for the drift tube other than water?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
All of the newer ELSDs can achieve excellent nebulisation with little temperature input from the nebulizer block or drift tube.A 40C temperature in the ELSD is not going to evaporate your solvents. Because you have water in your mobile phase, the tube temperature must be at least 120C.
With the nebulizer temperature off (or colder than the tube) you are creating a case where solvent can condense on the nebulizer.
This is most likely the cause of the noise (both factors combined).
Wow. They have changed that much over the past three to four years?All of the newer ELSDs can achieve excellent nebulisation with little temperature input from the nebulizer block or drift tube.