disapperance of peak
Posted: Mon Aug 27, 2007 2:21 am
by joyce
There were supposed to be 3 peaks on my standard, 1 internal standard and 2 other isomer, cis I and cis II. These 3 peaks appearded on my first run at 235C. However the third peak did not appear after the column oven temperature (230C) as well as when i set back to 235C.
Please help. Thank you.
Posted: Mon Aug 27, 2007 7:19 am
by k66
Give more details. what kind of substance. Did you make overheat befor right injetion?
Posted: Wed Aug 29, 2007 12:48 am
by joyce
Sorry for the brief statement.
My sample contains Cis I and Cis II isomer. Together with the internal standard, there are supposed to be 3 peaks, in which the retention time windows for the internal standard is apprximately 14 min; cis I- approximately 29 min; and cis II-approximately 27 min.
Most of the time when i only managed to get my cis I but never cis II. I have tried to adjust the oven temperature as well as the flow rate but it just doesn't seem to work.
I am using an autosampler with splitless injector and DB-1 capillary column.
Please help.
Posted: Thu Aug 30, 2007 2:04 am
by chromatographer1
Joyce,
Your information is still too sparse for the forum to help you.
Please try to clarify with more details: the analytes involved, the solvent used for injection, how much you are injecting, the injector and detector temperatures, the detector used, and the levels you are trying to detect.
Posted: Thu Aug 30, 2007 3:39 am
by joyce
I will try to explain as detail as possible.
I am using a 30m/0.25mm (id) with DB-1 of 0.25um film thickness coated with dimethyl polysiloxane packing.
The column temperature is between 225 to 235C, isothermal, injection prt is 260C and detector-300C, FID as rate is 0.8mL/min with the injection valume of 1uL.
I am using tetrahydrofuran as my solvent for my sample with dioctyl phthalate as my internal standard. I am supposed to test the percentage of the active ingredient in my sample, which would be around 10% (w/w).
Hopefully these details will be able to help me to solve this problem.