The issue I am having appears to be over-loading of the injection port. If I use the same conditions from the 5890 on the 7890 (flows, temperatures, splitless mode, etc.) the peaks are split on the 7890 output.
There are only 2 things that I have been able to do in order to improve peak shape; Turn on the split or inject less. Both of these solutions reduce the sensitivity of this analysis to the point that I can no longer use my usual low standard (1ppm MeOH, IPA and EtOH in water).
The problem persists if I change the flow type from constant fow to constant flow (like the 5890 would be).
I see there are many more options on the 7890 front panel, than the set of three knobs on the 5890.
General instrument conditions:
DB-WAXetr 30mx0.53 2u film
35°C (yes, this is below the recommended operating temperature) for 10 minutes, ramp to 200+
injection port 220
liner is a single taper split/splitless with quartz wool (part#5062-3587)
detector 250 (FID)
flow (have tried from 2-8 mL/min)
Samples are 10mL in 20mL vial, heated at 80°C for 20min. Syringe at 85°C. Have tried slow and fast injection speeds for the plunger.
I guess my question is: What is the 7890 doing differently to create this problem? I'm guessing the injection port's gas control is to blame, but what can I do?
One last thing I plan on trying soon: change the injection port liner to a smaller diameter splitless liner.