my hplc instrument gives a lot of fluctuation in back pressure, it sometimes increases to high pressure. what could be reason behind this and what is the solution for it?
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
More details please.
If you're running a gradient, this is quite normal. For isocratic runs, this signals a problem with the column or pumping system...(or the plumbing between them).
If you're running a gradient, this is quite normal. For isocratic runs, this signals a problem with the column or pumping system...(or the plumbing between them).
Thanks,
DR

DR

-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Or the detector flow cell
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
I think there is problem with the pump... please inspect the check valves.
4 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 28 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 28 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science