There is no analyte in the sample.
Peter Apps
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
Run a higher concentration sample and see what happens.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Run your standard again after the sample.
If the standard duplicates the first standard then there is no analyte in your sample as Peter has already stated, at least none present above your detection limit.
best wishes,
Rod
If the standard duplicates the first standard then there is no analyte in your sample as Peter has already stated, at least none present above your detection limit.
best wishes,
Rod
-
- Posts: 5
- Joined: Thu Aug 16, 2007 8:20 am
Thanks for the reply.
5 posts
Page 1 of 1
Who is online
In total there are 53 users online :: 1 registered, 0 hidden and 52 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 52 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 52 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science