one column 50*250mmL was packed with C4,10-15u,300A silica,the column efficiency was oK tested with Nathphthalene and phenanthrene
after that the column was plugged and stored at the lab,but ,one mother later,I retested the column efficiency,to my great surprise,I find that all the testing probe splitted?why is so ?then ,I reverse the column ,all the peak is normal!any suggestion will be highly appreciated!thank you!