why peak tailing

Posted: Fri Aug 17, 2007 6:12 am
by moon704520
hi all:

I am packing spring column 25mmID*250mmL,with 5u particle size,C18,but I find that the peak of the testing probe always tailing ,but I do not know why?can anybody help explain ?

Posted: Fri Aug 17, 2007 6:11 pm
by Mark Tracy
What is the test probe? There can be two reasons for tailing: 1) the stationary phase is damaged and a basic probe will be sensitive to that, or 2) the packing protocol is not optimized and a neutral probe will be sensitive to that. Have you talked to the manufacturer of the column hardware?

the testing probe is phenol and toluene ?and why tailing?

Posted: Mon Aug 20, 2007 7:20 am
by moon704520
Dear tracy:

the testing probe is phenol and toluene !the tailing factor is 1.3?I do not know why?is there any relationship bwtween the tailing factor and the flow rate?or there are something wrong with my packing protocol?do you have some valuable experience on the packing ?

Posted: Mon Aug 20, 2007 7:15 pm
by tom jupille
Actually, a tailing factor of 1.3 is not bad (I'm assuming the USP tailing factor measured at 5% of peak height). What have you achieved previously?

Posted: Mon Aug 20, 2007 10:21 pm
by Mark Tracy
Actually, for a neutral probe, a tailing factor of 1.3 is borderline, not terrible. First establish that the packing material is not damaged by looking a some basic probes. Second, budget yourself some time to empirically optimize the packing conditions: temperature, type of solvent, volume of solvent, weight of silica, compression rate and final pressure. Keep a notebook. When you have the optimized conditions, write an SOP. You will get better with practice.

Column manufacturers spend a lot of time developing this knowledge, and don't give it away.

would you please recommend me some acid and basic probes

Posted: Wed Aug 22, 2007 6:03 am
by moon704520
Dear tracy and Tom:


would you please recommend me some common used acid and basic probes?thank you very much!

Posted: Wed Aug 22, 2007 3:05 pm
by Mark Tracy
For basic probes, amitriptyline is a popular choice. We use it to test our columns. (80:20 MeOH:buffer pH 6.0 or 7.0) You should get an analytical-size column of the same type as your prep column to determine the expected behavior. Another popular basic probe is N,N-dimethylaniline.

For acid probes, 4-butylbenzoic acid is good.

You should also discuss this with the vendor of the silica you use. They will have QA tests; perhaps they will share them with you.

Posted: Wed Aug 22, 2007 8:22 pm
by tom jupille
Mark's suggestions are good.

However, if you are evaluating how well you have packed the column, a neutral probe is a better choice. Acidic or basic probes typically respond more to the chemistry of the column packing; tailing of neutral probes is usually related more to the quality of the packed bed.