Column Equilibration with ion pair reagent mobile phase

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Dear All
I am optimizing the method for analysis of some drug sample. As i am using ion pair reagent in mobile phase as compound is more polar. The method is isocratic method.
Now i am optimizing the ion pair regent concentration in mobile phase.
As the fact that, with ion pair regent mobile phase the column equilibration will take more time.
I am having one quiry as follow
Is the higher concentration of ion pair regent will take more time for column equilibration or low conc of the ion pair regent will take more time time for column equilibration ?


Thank you

The equilibrium condition for the ion-pair reagent is a fixed concentration of the reagent on the surface of the packing. While the surface concentration increases somewhat with the mobile phase concentration, the speed of equilibration increases a lot with the mobile phase concentration.
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 2 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry