Decrease of sensitivity at low concentration ethambutol.
Posted: Thu Aug 16, 2007 3:22 am
Dear All,
I have question about dramatical decreasing of peak area at low concentration.
I developed the LC/MS method to analyse ethambutol. I use Atlantis HILIC column elute with 75% acetonitrile (0.2% formic acid); isocratic system.
Peak area of ethambutol standard 1 mcg/ml is 4,000,000. Thus, I expect the peak area at 0.1 mcg/ml should be 400,000. However, the result is difference. The peak area at 0.1 mcg/ml is about 120,000.
Does anybody have suggestion?
Thank you,
Areeya