Re! Separation of Nitrogen and / or Oxygen from Laborato Air

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi!
We have a pharmaceutical Sample, which was formulated in an Nitrogen Atmosphere. We wanted to estimate the content of Nitrogen present in the formulated Vial. Can any body help me to solve my problem.
One of my colleauge suggested to use HP-PLOT Mole Sieve column at 40°C with TCD and Helium as Carrier.
Thanks in advance
Sreedhar/Ajith
+91 98661 89170 / 9849930916

Almost any 5A mole sieve column packed (or capillary,) 1 meter (to 30 meter), from any vendor, not just HP will suffice for the analysis.

Helium or hydrogen can be the carrier.

Temperatures of ambient to 60°C should be capable of separating nitrogen from methane/oxygen/CO/argon but not from krypton.

TCD is the appropriate detector.

hope this is helpful.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 139 users online :: 1 registered, 0 hidden and 138 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 138 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry