ECD incompatibility with Carbondisulphide

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello GC practitioners,

I have been trying for a week to seperate one and two carbon containing compounds dissolved in Carbondisulphide, and using an ECD as detector. I was seeing terrible tailing for CS2 and I just realised that ECD is not compatible with Carbondisulphide.

With the short literature search, all i could get was the info about the incompatibility, but neither the reason for it nor typical symptons that will be observed (I simply want to see if it matches with what i am observing). Would anybody have these information? Perhaps a list of compounds (used as solvents) incompatible with ECD (Other than chlorinated compounds offcourse!).

Thanks.

Shruesh,

I could not find exactly what you are looking for but did have the following:

"Detectors for Gas Chromatography" HP book by R. Buffington and M. K. Wilson,

"ECD response to various compounds
Hydrocarbons 1
Ethers, esters 10
Aliphatic alcohols, ketones 100
amines, mono CL, F cmpds
Mono-Br,di-Cl and di-F cmpds 1000
Anydrides and tri-Cl cmpds 10000
Mono-I, di-Br, poly-Cl and poly F 100000
Di-I, tri-Br, oly-Cl, poly-F 1000000

Unfortunately no mention of your problem, sulfur compounds which also typically respond, see Pick, M.E., Jrnl of Chrom. 171, (1979), p. 305. Also no mention of nitro compounds which also respond very well (search ECD and explosives) or oxygenates. Also search Pelilizzari, E.D. and electron capture (afraid I don't have the full reference.)


Sorry not more helpful...

Best regards.

It's many years ago now but I remember a colleague measuring carbon disulfide in toluene by ECD. In a splitless injection I think the sensitivity was just about enough to detect 0.02 % w/v in toluene which may not be good for your application.
was seeing terrible tailing for CS2 and I just realised that ECD is not compatible with Carbondisulphide.
hmm Strange and inexplicable behavior :)
What's the problem? It's plain to give strong signal.
In my laboratory we are using uECD to determination CS2 in food without any problems. It's correct detector to detarmination CS2 but u can't use CS2 as a solvent or carry gas.
4 posts Page 1 of 1

Who is online

In total there are 55 users online :: 1 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Georguslat and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry