Contamination from ion-pairing agents.
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
I used the tributylamine in mobile phase as an ion-pairining agent for the retention of acids on revered phase column. After doing lot off washing with different solvents, still i observed the peak at 186, which is definately from protonated tributylamine. I have also changed column, but the peak is there. Can anybody give the some useful tips to remove the contamination of tributylamine from the instrument.
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
At one point I had an HPLC that that I used for glyphosate by post column derivatization and on occasion, diquat. The diquat separation required diethyl amine in the mobile phase while the post column reaction converted amines to a fluorescent iso-indole. When the system was new I could make the conversion without difficulty with about a day of flushing and cleaning. After 4 years the cleaning process involved much more extensive dismantling and even replacement. I finally started borrowing a different HPLC to run the diquats, it was that bad.
The amine can reside on any surface the mobile phase contacts. Some are easy to wash off; tubing, pump heads, but others are harder to get to; e.g. sampling valve faces. I found that taking apart the sampling valve for cleaning saved a lot of grief, as did changing all frits in the system.
The amine can reside on any surface the mobile phase contacts. Some are easy to wash off; tubing, pump heads, but others are harder to get to; e.g. sampling valve faces. I found that taking apart the sampling valve for cleaning saved a lot of grief, as did changing all frits in the system.
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
Have a look in one of the previous posts:
http://www.sepsci.com/chromforum/viewto ... ight=rapid
As you can see triethylamine can be a huge problem when used with mass spectrometry (especially ion traps) and I would expect you to have similar problems from tributylamine
http://www.sepsci.com/chromforum/viewto ... ight=rapid
As you can see triethylamine can be a huge problem when used with mass spectrometry (especially ion traps) and I would expect you to have similar problems from tributylamine
3 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 1 guest
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science