Symmetrical changes in spectrum during Peak elution
Posted: Fri Aug 10, 2007 12:41 pm
In an HPLC method I am currently developing I am observing an odd phenomenon:
The peak of the active as such is perfectly symmetrical (USP Tailing 1.04).
However, when scanning over the peak, recorded with a PDA detector in the range 190-400 nm, the spectrum changes quite significantly in the wavelength region up to approx. 250 nm. At the beginning and end of the peak, the spectrum would have a maximum absorbance at approx 200nm, whereas in the centre of the peak, the maximum of the band is 242nm.
I am quantifying the peak at 301nm, a maximum of another absorption band which is unchanged during the elution of the peak.
The molecule contains an indole system, a sulfonamide structure, and a carbamate structure. The mobile phase is acetonitrile/ammonium formiate buffer pH 2.8. The standard solution contains additionally, to support dissolution, 10% DMF.
Could anyone explain the symmetrical changes in the spectra throughout the peak elution, also there is no indication that it would be a mixture of different compounds?
Thank you very much!