GC/MS P&T - High CO2 - Low compound recovery

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
HI,
I'm running a GC/MS P&T method using a 6890N, 5975B and a velocity XPT. When I run fresh water samples with a volume of 5ml i get a large CO2 peak and a low target analyte recovery. When I run QC, purged H2O, samples or soil samples, extracted in MeOH, I get a smaller hit for CO2 and analyte recovery is fine.
Because I am not getting large CO2 hits for all samples, I don't think it is a problem with the trap.

Can anyone help me?

cheers

There is CO2 desolved in almost all water samples, it's no surprise to see a big CO2 peak in the front. On the other hand, there is little CO2 in the purged water, CO2 peak is much smaller in your QC samples. CO2 has nothing to do with your problem. If QC is ok and you only have problem with matrix spike, it could be matrix interference or the way how you prepare the sample spike solution. Can't comment further without detail information.
2 posts Page 1 of 1

Who is online

In total there are 124 users online :: 1 registered, 0 hidden and 123 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 123 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry