Reproducibility
Posted: Mon Jul 30, 2007 8:29 pm
I'm running a Berger SFC but it's a DAD Agilent detector so I figured this question would get more input this way...
I ran the same sample for time (sample containing two compounds);
Peak 1 Peak 2 Ratio
1 1237 3663.4 0.33766446
2 887.3 2674.1 0.33181257
3 847.5 2520.6 0.33622947
4 791.1 2348 0.33692504
RSD (%) = 0.78
So my RSD is fairly good (not surprising since I'm considering the ratios)... But are the numbers themselves too different from one run to another? I mean, we are using an autosampler (Alcott), with a fixed loop system (loop is 5 uL)... Is it normal that the areas vary so differently from run to run?
Thanks!
Trishia.
I ran the same sample for time (sample containing two compounds);
Peak 1 Peak 2 Ratio
1 1237 3663.4 0.33766446
2 887.3 2674.1 0.33181257
3 847.5 2520.6 0.33622947
4 791.1 2348 0.33692504
RSD (%) = 0.78
So my RSD is fairly good (not surprising since I'm considering the ratios)... But are the numbers themselves too different from one run to another? I mean, we are using an autosampler (Alcott), with a fixed loop system (loop is 5 uL)... Is it normal that the areas vary so differently from run to run?