Just a quick question...

I seem to remember that when we did injections on the GC in my previous lab, we would push the needle all the way into the injection port (until the glass syringe hit the metal)...

In my current lab, we use a CTC-PAL autosampler to do the injections (Agilent 6850 GC)... The needle only goes partly in by default - is this a concern for reproducibility or because we're using an autosampler it's completely irrelevant...?


Thanks!

Roxanne.

Roxanne42,

Autosamplers are typically very reproducible on sampling depth so this should be irrelavent. On the 5890's there was a heat profile for the injector so sampling depth would affect discrimination. I suspect the same may be true for 6850's so you might try different heights if you need to alter sensitivity or discrimination.

Syringe all the way to the metal nut is also very reproducible, especially for manual injections, which is why I suspect people do it that way.

Best regards.
2 posts Page 1 of 1

Who is online

In total there are 84 users online :: 1 registered, 0 hidden and 83 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 83 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry