Crude oil analysis by GC [August 29, 2004]
Posted: Mon Aug 30, 2004 8:00 pm
by admin
By naderih on Sunday, August 29, 2004 - 06:49 am:
Is there any method for analysis of crude oil by GC directly? It is possible to inject the oil to GC.
Posted: Thu Sep 02, 2004 7:13 pm
by Davide
What kind of oil do you mean ? Minerall or vegetal ?
Posted: Tue Sep 07, 2004 12:06 pm
by emsman
depending on the components you might have make a derivatisation before injecting on the GC. For example fatty acids are to be esterified, but then You can detect them up to at least C36.
Posted: Tue Sep 07, 2004 6:04 pm
by Davide
I have experience in the analysis of free fatty acid. You can analyse it without derivatisation in water sample if you use a DB-FFAP column with MS. If the amount of acid is a low level use a splitless injection mode
Crude Oil
Posted: Tue Sep 21, 2004 8:46 pm
by semivoadude
try extracting the oil 0.1g to 10 ml of Dichloromethane, then inject 2ul. You will see the entire C9-C40+ range
Posted: Mon Oct 04, 2004 3:57 am
by Ron
I assume you mean petroleum. Look in any column catalog or web site. You will find chromatograms and conditions.