If you are required to follow SW-846; BFB or DFTPP are required for full scan analysis. For example section 9.3 of 8270D:
9.3 Quality control procedures necessary to evaluate the GC system operation are
found in Method 8000 and include evaluation of retention time windows, calibration verification
and chromatographic analysis of samples. In addition, discussions regarding the instrument QC
requirements listed below can be found in the referenced sections of this method:
• The GC/MS must be tuned to meet the recommended DFTPP criteria prior to the
initial calibration and for each 12-hr period during which analyses are performed.
See Secs. 11.3.1 and 11.4.1 for further details.
The new versions, 8260C and 8270D have more expicit allowance of alternate tune criteria.
For answers to SW-846 method questions contact the MICE service at:
http://www.epa.gov/epaoswer/hazwaste/test/mice.htm
They are paid to answer your questions!