Can anybody help me with troubleshooting on a 6890 GC? I did a maintainance last week and don't get it running again since then. I changed liner, o-ring, gold seal, column ferrule and split vent line. Problem was then, that I didn't reach a pressure higher then 90 kPa. I checked all off the changed parts again. It got better so far, that I reach a higher pressure, but only as long as gas saver mode is on. I can start the sequence and instrument gets the sample, but just a moment before injection the pressure breaks down. This because the gas saver mode is turned off for injection. If I make a splitless method, then there is no problem. Can there be a problem with the EPC-module? If there is a leak, splitless mode should also fail, doesn't it?