Page 1 of 1

sugar sugar sugar

Posted: Mon Jul 23, 2007 8:48 pm
by srdales
Any advice on the best wash procedure to get sugar syrup out of the HPLC system?


HELP - this is not a SWEET issue. :shock:

Posted: Mon Jul 23, 2007 10:11 pm
by juddc
Have you tried warm water?

Posted: Tue Jul 24, 2007 12:43 pm
by srdales
Yes, we have tried hot water, but it does not remove it all.

Posted: Tue Jul 24, 2007 1:10 pm
by DR
Keep pumping. You'll get it eventually. Also - load some sample vials w/ water & do frequent, large injections to get any of it out of the injector plumbing too.

Posted: Tue Jul 24, 2007 5:32 pm
by Kostas Petritis
Let me ask this, did you just injected large amounts of sugar syrup or did you try to pump it? Maybe I shouldn't be asking that, but you never know...

If you just injected it, you can dissassemble your injection valve and sonicate the rotor etc...

Posted: Tue Jul 24, 2007 11:05 pm
by tom jupille
where in the system? injector needle? injection valve?, column?

Posted: Wed Jul 25, 2007 1:03 am
by srdales
The injections are of oral drug syrups that are sucrose based. Due to the impurities quantitation, larger concentration must be injected.

The samples are diluted down with appropriate solvents. Straight sucrose is not being injected.

We have large number of oral drug syrup products that are sucrose based. The use of the systems are showing great signs of carryover.

We are looking for recommendation for the injector wash and the system wash.

I appreciate all the assistance on this. Any suggestion is appreciated.