We just got a new Acella LC system (just the LC, no MS) and have been tryingh to hook it up ourselves (long story). Anyway, we can get the pump to talk to the software (its controlled via USB) but the autosampler and PDA are ethernet connected going through a switch box. We just cant get them to talk to the software (or be seen by the software). I suspect this is more of a Windoze XP problem then a problem with the components or the software, anyone out there have any words of advice (calling Thermo/Fisher isnt an option, yet, again, long story).

Here's hoping.....

AA

What IP have you set your Lan Connection to? It should be 172.16.0.101 w/ subnet mask 255.255.0.0
Kind regards
Leadazide

Thank you, that made it work.

No problem...

Just out of curiosity... what has gone wrong with local Thermo department? And what area is it, just for future refernce.. :)
Kind regards
Leadazide
4 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Bing [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry