Is 0.02 seconds for each delay setting for a 2 channel MRM method with a 0.5 second dwell time typical? Is that too fast for the timers to refresh and reset?
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Would you mind elaborating on that a bit?Cycle time can affect your quantitation accuracy, dwell times can affect your sensitivity. New generation triple quadrupoles can go pretty low dwell times (i.e. 25-50 ms) without minimal effects on sensitivity. They are also much more resistant to cross-talking...
Again for older instruments and for mutliple component analysis, if the chromatography allowed it (i.e. if there was a part in the middle of the chromatogram without any peaks) you could create two events with different transitions (e.g. 10 transitions followed in the first part and 12 transitions on the second). In this way you could afford higher dwell times...