Based on recent information during an LC/GC webinar on AIQ, the presenter mentioned that you should have "global" type procedures for calibration so all instruments (HPLC) are being calibrated in a similar manner.
My question pertains to detector linearity. How does your company perform detector linearity, and does the procedure account for the different absorbance ranges of the different manufacturer models?
Does anyone simply use the ASTM procedure by running 5 standards, then calculating response factor? (Requirements being the rsd of the response factors must be less than 5%) Or does anyone simply calculate slope and intercept?
I'm just curious if there is an industry standard. Thanks.
Russell
My question pertains to detector linearity. How does your company perform detector linearity, and does the procedure account for the different absorbance ranges of the different manufacturer models?
Does anyone simply use the ASTM procedure by running 5 standards, then calculating response factor? (Requirements being the rsd of the response factors must be less than 5%) Or does anyone simply calculate slope and intercept?
I'm just curious if there is an industry standard. Thanks.