Chromatogram spikes

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear all

good morning

I am doing an analysis of methyl acetate on an Agilent 6890 using Helium as carrier gas (FID detector) using the GC injector. I am observing continuous spikes in the chromatogram. I have baked the column overnight, checked the inlet septa, cleaned the jet plug and ensured there are no leakages. What could be the reason and how can i avoid it ?
Other GC instruments connected to the same gas lines do not give this problem. It is specific to one particular instrument, the one that i am using.

Thanks and regards
Dinesh

Are the spikes still there if you do a blank run with no injection ?

Are the spikes there during standby ?

Are the spikes there if you turn the flame off. ?

Remove the column from the detector and plug the connection - are the spikes still there ?

Peter
Peter Apps

What kind of column are you using?

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Prabhu1 and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry