Dear members,
I always use the value of Rs not less than 2 during method development for simple mixture. This value is hard to get when we use more complex mixture (containing 10 or more components). Is it Rs = 1 or 1.5 enough in this case? Thanks in advance.

Regards,
Siswanto

Yes, in cases where it is difficult to get resolution, it will be ok.

I have seen a warning letter, in which the resolution limit was 1 in their STP but their method was capable of giving higher resolution. So the point of USFDA was, the limit should be higher as you method of capable of achieving that.

Regards
jUST dO iT....

As far as I know, Rs > 2 is a "recommendation" in the FDA "Guidance for Reviewers", not a requirement. If your validation results for accuracy and precision are otherwise acceptable, a lower resolution may be OK (and, conversely, a higher resolution may be required!).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry