- Posts: 2
- Joined: Tue Dec 20, 2005 4:02 pm
When i'm doing my analysis i got a constant drifting baseline. It's a straight line that's going up. I already purged the RI and the pump several times, and refreshed the solvent cyclohexane. Still my baseline is going up. I checked the RI temperature, and the average temp. is 39.985°C +/- 0.005 (small peaks till 40.0 and 39.975), the signal of the reference cell is constant, the other cell is constantly going down.. I use a flow of 0.4ml/min.
The column i use i made myself and exist out of a freshly calcinated zeolite (so all dirt is gone), wich i packed under N2 atmosphere. The poresize of the zeolite is choosen so that the solvent can't enter the pores. I'm already flushing for 4 days, but still no change.














 
																							 .
. 