Hi at all,
I have another problem:
I don't understand why some time I have the double peak of same molecule. I analize aromatic amine.
Thanks
Antonio

Without knowing the exact structure we can give you only general answers. Isn't it possible that molecule may have more enantiomers?

This can happen if the pK of the analyte is close to the pH of the mobile phase and the buffering capacity of the mobile phase is low and the sample pH is mismatched. If this is happening in your case you can change the pH of the mobile phase so that it is farther away from the pk of your analyte, you can increase the buffering capacity of your mobile phase, or you can prepare the sample in mobile phase.

It can also happen if your compound has two isomers in a slow equilibrium.
Mark Tracy
Senior Chemist
Dionex Corp.

1, bad column;
2, over loaded column;
3, strong diluent;
4, bad connection;
5, degradation;
6, contamination;
7, low quality mobile phase if gradient
Excel
4 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry