I'm trying to troubleshoot a GC-ECD with N2 carrier set up to run N2O using porapak Q columns. I think I have a leak but want to check where and what the leak is using a continuous flow isotope mass spec and Argon. In order to do this I will need to run He rather than N2 as the carrier. Will this pose a problem for the GC or columns, or subsequent analysis when I switch back to N2 as the carrier? Would 24 hours be sufficient purging of He?

Thanks in advance,
Matt

Matt,

Switching to helium while you hunt for a leak should not harm anything. Your ECD might be unhappy since there is no beta moderator but you could turn it off while you find your leak. You might need to let you column re-equilibrate after you switch back but it should not take very long.

By the way, what levels of N2O are you looking at with the ECD?

Best regards.

Thanks for the reply. The GC is purging with He as I write this. I've capped off the ECD and will run the tests with it off, thanks. We are looking at N2O concentrations typically around 300ppb+.
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry