HPLC PDA and wavelength ratios
Posted: Thu Jun 28, 2007 1:20 pm
I recently evaluated analyses on an HPLC with an HPLC photodiode array detector that showed a systematic increase in the peak area ratio of 664/450 nm when the ratio was plotted as a function of the 450 peak area. These peak areas were obtained during analysis of multi-level chlorophyll a calibration standards and for which there were no co-elution problems, ghost peaks, and SNRs were adequate. This phenomenon was observed on a ThermoQuest detector using a flow cell with a 5 cm pathlength. I observe a constant ratio on my Agilent 1100 photodiode array detector as does another colleague of mine who has a Shimadzu (both of us have 1 cm flowcells) when these standards are analyzed. I am especially interested in this because the analyst with the ThermoQuest HPLC experiences non-linear calibration curves, where as I do not. I would really like to know if this changing ratio could be symptomatic of an inability to obtain a linear calibration. Any thoughts? (P.S. Chlorophyll a absorbs strongly at 450 and 664 and the standards used covered a wide concentration range).