SRI8610 GC. Volatile fatty acids(VFA) detection
Posted: Wed Jun 27, 2007 7:37 pm
Following are the operating conditions:
GC-SRI 8610
Carrier: Helium
Detector:FID
Column: HAYSEPQ 80/100 Packed column
Temperatures: Oven 200 C; Detector:220 C;
I was having good peak areas, in magnitude of 12-15000 for Acetate;propionate;butyrate;
I changed septa, and surprisingly, magnitude of the peak areas reduced immediately by 5 times, to 2000.
I checked for all the leaks, changed the syringe, checked the liner connection,
It is very surprising, that peak areas have dropped in fraction of minute; after septum change.
Any ideas please????
Ramana
GC-SRI 8610
Carrier: Helium
Detector:FID
Column: HAYSEPQ 80/100 Packed column
Temperatures: Oven 200 C; Detector:220 C;
I was having good peak areas, in magnitude of 12-15000 for Acetate;propionate;butyrate;
I changed septa, and surprisingly, magnitude of the peak areas reduced immediately by 5 times, to 2000.
I checked for all the leaks, changed the syringe, checked the liner connection,
It is very surprising, that peak areas have dropped in fraction of minute; after septum change.
Any ideas please????
Ramana