Please advise how to interfase a Tekmar P&T Aquatek 70 to an Agilent 6890N/5975B system so that the P&T start the GC/MSD automatically.
-
- Posts: 303
- Joined: Fri Sep 03, 2004 8:06 pm
I have two HT3 purge+trap instruments, and Tekmar supplied us with a cable that connects the autosampler to the GC. Perhaps give them a call and see if they can get a cable to you.
-
- Posts: 196
- Joined: Mon Jan 29, 2007 10:19 pm
Hi,
the following may help you connect injector to the 6890, It was used to connect a PE ATD to the 6890, you need to refer to the autosamplers' manufacturer for cable connections.
The HP6890 GC uses one TTL connection for both START IN and START OUT and uses a second TTL connection for READY IN and READY OUT.
The HP6890 GC does not have a READY OUTPUT RELAY or a START OUTPUT RELAY and, therefore, it needs a special cable to communicate with other instruments in the system.
the instruction manual can be gotten at
http://chemstation.googlepages.com/PESa ... 100410.pdf
and you can purchase the cable from Perkin Elmer : P/N N610-0410
regards
Alex
the following may help you connect injector to the 6890, It was used to connect a PE ATD to the 6890, you need to refer to the autosamplers' manufacturer for cable connections.
The HP6890 GC uses one TTL connection for both START IN and START OUT and uses a second TTL connection for READY IN and READY OUT.
The HP6890 GC does not have a READY OUTPUT RELAY or a START OUTPUT RELAY and, therefore, it needs a special cable to communicate with other instruments in the system.
the instruction manual can be gotten at
http://chemstation.googlepages.com/PESa ... 100410.pdf
and you can purchase the cable from Perkin Elmer : P/N N610-0410
regards
Alex
3 posts
Page 1 of 1
Who is online
In total there are 10 users online :: 1 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 9 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 9 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science