extra column volume
I would like to ask if it is possible to practicly measure the extra column volume of my chromatographic system. I have alrady tried it, without using the column. I found that the extra column volume was too high to be correct. Can anybody find a logical way of measuring these volumes in the laboratory?
Thanks for any future help
The extra-column volume, i.e. the volume between the injector and the detector, can be measured by connecting the injector outlet tubing directly to the detector inlet tubing, and injecting a small volume of any analyte, for example acetone dissolved in mobile phase. Considering that such a volume will be rather small, it is best to do this experiment at slow flow rates with a high detector sampling rate. For a standard HPLC, I would do this at around 0.3 mL/min, and the expected value is around 100 microliter. Under these circumstances, your injection volume should be less than 2 microliter.
Thank you very much for your advice and help!
Another option is to leave the column in place and run a series of standards (isocratically). Plot σ^2 (σ is 1/4 of the peak baseline width) versus tr^2. The y-intercept gives an estimate of the extra-column contribution to σ^2.
Look at slide 8 from this paper for details:
http://www.lcresources.com/resources/ex ... 060223.pdf
Thank you very much for your advise...I think this is a good way to demonstrate it in theory
